Yeah, I was just ssh-ing into the Python code and sudo saving changes. This approach would make it a lot easier. I will do that.
Well I wasn’t sure what was causing it when it first happened. The culture is visibly very dense and I didnt have the voltage readout of the 90 degree PD because I was using REF so I only see the resulting OD from dividing them.
Yes, you understand correctly. These were the conditions when the calculated offset value was larger than the initial OD readings at a low culture density.
Should we not expect a larger voltage value seeing as the offset is calculated at 0 light condition which is similar to approaching a very high OD (at least for the 90 PD)?
The test is being run with channel 1 REF OD just looking at light source and channel 2 OD has a very dense culture between it and light source. Here are the debug logs for the 2 test conditions.
1. Code running as intended (light off during offset calculation)
2024-02-01T23:54:37-0500 DEBUG [ir_led_ref] Using PD channel 1 as IR LED reference.
2024-02-01T23:54:37-0500 DEBUG [calibration_transformer] No calibration available for channel 2, angle 90, skipping.
2024-02-01T23:54:37-0500 DEBUG [od_reading] Init.
2024-02-01T23:54:38-0500 WARNING [od_reading] The value for the IR LED, 100.0%, is very high. We suggest a value 90% or less to avoid damaging the LED.
2024-02-01T23:54:38-0500 DEBUG [od_reading] Starting od_reading with PD channels {'2': '90'}, with IR LED intensity 100.0% from channel A.
2024-02-01T23:54:41-0500 DEBUG [adc_reader] Using ADC class Pico_ADC.
2024-02-01T23:54:41-0500 DEBUG [adc_reader] ADC ready to read from PD channels 2, 1, with gain 1.
2024-02-01T23:54:46-0500 DEBUG [adc_reader] AC hz estimate: 60.0
2024-02-01T23:54:46-0500 DEBUG [adc_reader] timestamps={'2': [4.890599666396156e-05, 0.02764456099976087, 0.05594922099407995, 0.08386023799539544, 0.11246583399770316, 0.14059143299527932, 0.1682427639971138, 0.19665804899705108, 0.2246074509967002, 0.2533061189969885, 0.2815366649956559, 0.30931940199661767, 0.3378260399986175, 0.3658659619977698, 0.39344896099646576, 0.42180674599512713, 0.4496868770002038, 0.4783224729981157, 0.5064966129939421, 0.5342114329978358, 0.5626479159982409, 0.5906183069964754, 0.6193735380002181, 0.6476465829982772, 0.6754541639966192, 0.7040202279968071, 0.7321233269976801, 0.7597524189986871, 0.7881212979991687, 0.8160710639931494, 0.844745669994154, 0.8729459039968788], '1': [0.0013934809976490214, 0.02880637599446345, 0.05713640099565964, 0.0851355419945321, 0.11369197199383052, 0.14169585299532628, 0.16932088199973805, 0.19783215599454707, 0.22571020399482222, 0.25445439299801365, 0.28262749099667417, 0.3105079359957017, 0.33890118899580557, 0.3669487679944723, 0.39451765199919464, 0.42301559199404437, 0.45079989099758677, 0.47947199699410703, 0.5076615539946943, 0.5353776759948232, 0.5637186899984954, 0.5916923629993107, 0.6205807689984795, 0.6487602219931432, 0.6765777499967953, 0.7052088669952354, 0.7332392579992302, 0.7608309539937181, 0.7892389469998307, 0.8172773589976714, 0.8459017039931496, 0.8740285530002438]}
2024-02-01T23:54:46-0500 DEBUG [adc_reader] aggregated_signals={'2': [1419, 1573, 901, 1199, 1853, 1156, 1070, 1837, 1292, 818, 1644, 1539, 807, 1455, 1618, 842, 1344, 1826, 1038, 1146, 1850, 1182, 864, 1712, 1465, 807, 1503, 1597, 820, 1360, 1821, 1002], '1': [1328, 1428, 1296, 1294, 1443, 1341, 1263, 1430, 1360, 1240, 1371, 1414, 1248, 1335, 1438, 1279, 1319, 1453, 1328, 1290, 1443, 1349, 1245, 1391, 1403, 1248, 1344, 1426, 1273, 1323, 1453, 1314]}
2024-02-01T23:54:47-0500 DEBUG [adc_reader] timestamps={'2': [5.677100125467405e-05, 0.027667684997140896, 0.05604750099882949, 0.08392893499694765, 0.11251255199749721, 0.1406446609980776, 0.1683021899953019, 0.19669148499815492, 0.22464791799575323, 0.25335106599959545, 0.281580048998876, 0.3094045039979392, 0.33792379799706396, 0.3659771059974446, 0.39356177100125933, 0.42189325399522204, 0.44978156299475813, 0.47841403300117236, 0.5065723919979064, 0.5342785149987321, 0.5627167679995182, 0.590691899000376, 0.6194100979992072, 0.6476786119965254, 0.6754947340014041, 0.7040808509991621, 0.7321782209983212, 0.7598298640004941, 0.7882052529967041, 0.8161202799965395, 0.8447934279975016, 0.8730031929953839], '1': [0.002031288997386582, 0.028929916996276006, 0.05734952399507165, 0.08518481200007955, 0.11365160699642729, 0.14182975799485575, 0.16938285999640357, 0.19777324899769155, 0.22585218099993654, 0.25454991199512733, 0.28268332299921894, 0.31066376699891407, 0.33915540499583585, 0.36706480699649546, 0.39465108700096607, 0.42299772599653807, 0.45097743999940576, 0.4795378799972241, 0.5076756139969802, 0.5354480899986811, 0.5638062919970253, 0.5917699130004621, 0.620495454997581, 0.6488330840002163, 0.6766116539947689, 0.7052836029979517, 0.7333251919990289, 0.7610321479951381, 0.7892863919987576, 0.817231314998935, 0.8460059200006071, 0.8741328210016945]}
2024-02-01T23:54:47-0500 DEBUG [adc_reader] aggregated_signals={'2': [1091, 1811, 1059, 1104, 1837, 1290, 991, 1792, 1413, 816, 1537, 1615, 848, 1329, 1715, 912, 1232, 1840, 1152, 1064, 1826, 1299, 832, 1616, 1551, 822, 1378, 1685, 873, 1282, 1837, 1118], '1': [1276, 1451, 1329, 1281, 1440, 1364, 1249, 1419, 1390, 1248, 1347, 1436, 1280, 1315, 1447, 1296, 1303, 1452, 1341, 1273, 1433, 1367, 1245, 1361, 1415, 1266, 1327, 1442, 1289, 1310, 1448, 1342]}
2024-02-01T23:54:47-0500 DEBUG [adc_reader] ADC offsets: {'2': 1326, '1': 1348}, and in voltage: {'2': 0.06678571428571428, '1': 0.0678937728937729}
2024-02-01T23:54:50-0500 INFO [od_reading] Ready.
2024-02-01T23:54:50-0500 DEBUG [od_reading] od_reading is blocking until disconnected.
2024-02-01T23:54:51-0500 DEBUG [adc_reader] timestamps={'2': [4.9531001423019916e-05, 0.02771018500061473, 0.056005001002631616, 0.08385560100578004, 0.112422135003726, 0.1405287750021671, 0.1682229180005379, 0.1966374210023787, 0.22473130100115668, 0.2536617910009227, 0.28197139800613513, 0.309828509001818, 0.33844545900501544, 0.36659001600492047, 0.39420869100285927, 0.4225568920010119, 0.4504862419998972, 0.47923412800446386, 0.5073360810056329, 0.5350363180041313, 0.5634945180063369, 0.5915351700023166, 0.620286388999375, 0.6485465180012397, 0.6763648270061822, 0.7049393810011679, 0.7330213860041113, 0.7606626130000222, 0.7890500850044191, 0.8170496950042434, 0.8456797170001664, 0.8738542740029516], '1': [0.0020454550030990504, 0.02887496900075348, 0.05714978500327561, 0.08501288600382395, 0.11350124200544087, 0.14160606000223197, 0.16947265000635525, 0.19778251800016733, 0.22608483400108526, 0.2549642820013105, 0.2832686300025671, 0.3111477720012772, 0.3397878980031237, 0.368072558005224, 0.3954849319998175, 0.42372521800280083, 0.4518133689998649, 0.4804622970041237, 0.5084962300024927, 0.536129123000137, 0.5648250830054167, 0.5927097450039582, 0.6214295590034453, 0.6496786980060278, 0.6775702880040626, 0.7060402070055716, 0.7341062230043462, 0.7618543240023428, 0.7902492440043716, 0.81832958200539, 0.8467743460059864, 0.8750594740049564]}
2024-02-01T23:54:51-0500 DEBUG [adc_reader] aggregated_signals={'2': [1369, 1319, 1358, 1359, 1438, 1420, 1308, 1472, 1376, 1306, 1370, 1217, 1328, 1350, 1328, 1364, 1361, 1464, 1431, 1295, 1456, 1345, 1303, 1373, 1216, 1329, 1338, 1308, 1360, 1358, 1447, 1425], '1': [981, 1015, 996, 982, 1023, 992, 980, 1008, 975, 992, 990, 993, 993, 980, 1013, 997, 978, 1018, 987, 984, 997, 973, 988, 987, 988, 992, 988, 1012, 995, 983, 1021, 986]}
2024-02-01T23:54:51-0500 DEBUG [od_reading] IR Reference is 0.0. Is it connected correctly? Is the IR LED working?
File "/usr/local/lib/python3.11/dist-packages/pioreactor/utils/timing.py", line 147, in _execute_function
self.function(*self.args, **self.kwargs)
File "/usr/local/lib/python3.11/dist-packages/pioreactor/background_jobs/od_reading.py", line 983, in record_from_adc
od_reading_by_channel = self._read_from_adc_and_transform()
File "/usr/local/lib/python3.11/dist-packages/pioreactor/background_jobs/od_reading.py", line 1088, in _read_from_adc_and_transform
return self.calibration_transformer(self.ir_led_reference_tracker(batched_readings))
File "/usr/local/lib/python3.11/dist-packages/pioreactor/background_jobs/od_reading.py", line 619, in __call__
return {
File "/usr/local/lib/python3.11/dist-packages/pioreactor/background_jobs/od_reading.py", line 620, in <dictcomp>
ch: self.transform(od_signal)
File "/usr/local/lib/python3.11/dist-packages/pioreactor/background_jobs/od_reading.py", line 629, in transform
raise ValueError("IR Reference is 0.0. Is it connected correctly? Is the IR LED working?")
ValueError: IR Reference is 0.0. Is it connected correctly? Is the IR LED working?
2024-02-01T23:54:51-0500 ERROR [od_reading] IR Reference is 0.0. Is it connected correctly? Is the IR LED working?
2024-02-01T23:54:56-0500 DEBUG [adc_reader] timestamps={'2': [4.713499947683886e-05, 0.027695600998413283, 0.055988385996897705, 0.08382195499871159, 0.1123928109955159, 0.14052356599859195, 0.1681851569956052, 0.1965837219977402, 0.22454197800107067, 0.2533318429996143, 0.281675460995757, 0.3096709559977171, 0.3383085300010862, 0.36684933400101727, 0.3945273829958751, 0.42295245899731526, 0.4509273809962906, 0.47962693400040735, 0.5078013859965722, 0.535504695995769, 0.5639425839981413, 0.5919405789973098, 0.6206763289956143, 0.6489429159992142, 0.6767517459957162, 0.7053028629961773, 0.7333940339958644, 0.7610254169994732, 0.7893960660003358, 0.817308124002011, 0.8459958539970103, 0.8742442639995716], '1': [0.0014066049989196472, 0.028997779998462647, 0.05713754500175128, 0.08494127100129845, 0.11352144999545999, 0.14175246399827302, 0.16932363899832126, 0.19769725699734408, 0.22577222999825608, 0.25466079299803823, 0.2831437319982797, 0.31143422499735607, 0.33962357299606083, 0.36819437699887203, 0.395925446995534, 0.4243956790014636, 0.4522455499973148, 0.4809048419992905, 0.5089588789996924, 0.536657552998804, 0.5650239309979952, 0.5930783840012737, 0.6218158529954962, 0.6501029610008118, 0.6778426759992726, 0.7063909799981047, 0.7345473079985823, 0.7621179099951405, 0.790494861001207, 0.8183963459960069, 0.8472348040013458, 0.8755020159951528]}
2024-02-01T23:54:56-0500 DEBUG [adc_reader] aggregated_signals={'2': [1338, 1337, 1346, 1358, 1378, 1314, 1447, 1427, 1280, 1444, 1255, 1297, 1336, 1354, 1350, 1359, 1432, 1424, 1281, 1472, 1381, 1287, 1406, 1217, 1297, 1328, 1331, 1343, 1360, 1403, 1412, 1281], '1': [1014, 994, 986, 1015, 998, 976, 1019, 984, 982, 992, 986, 991, 990, 1013, 993, 977, 1018, 981, 985, 1012, 976, 986, 987, 988, 990, 990, 1007, 994, 982, 1022, 992, 979]}
2024-02-01T23:54:56-0500 DEBUG [od_reading] IR Reference is 0.0. Is it connected correctly? Is the IR LED working?
return {
2024-02-01T23:54:56-0500 ERROR [od_reading] IR Reference is 0.0. Is it connected correctly? Is the IR LED working?```
2024-02-02T00:13:22-0500 DEBUG [ir_led_ref] Using PD channel 1 as IR LED reference.
2024-02-02T00:13:22-0500 DEBUG [calibration_transformer] No calibration available for channel 2, angle 90, skipping.
2024-02-02T00:13:22-0500 DEBUG [od_reading] Init.
2024-02-02T00:13:22-0500 WARNING [od_reading] The value for the IR LED, 100.0%, is very high. We suggest a value 90% or less to avoid damaging the LED.
2024-02-02T00:13:22-0500 DEBUG [od_reading] Starting od_reading with PD channels {'2': '90'}, with IR LED intensity 100.0% from channel A.
2024-02-02T00:13:25-0500 DEBUG [adc_reader] Using ADC class Pico_ADC.
2024-02-02T00:13:26-0500 DEBUG [adc_reader] ADC ready to read from PD channels 2, 1, with gain 1.
2024-02-02T00:13:30-0500 DEBUG [adc_reader] AC hz estimate: 60.0
2024-02-02T00:13:30-0500 DEBUG [adc_reader] timestamps={'2': [5.0677001127041876e-05, 0.027640685002552345, 0.055923292005900294, 0.08378725700458745, 0.11236449700663798, 0.1404729390051216, 0.16812836300232448, 0.1965605530058383, 0.22450857900548726, 0.2531877460060059, 0.28141207300359383, 0.309173642002861, 0.3377399970049737, 0.365785471003619, 0.39338756200595526, 0.4217233990057139, 0.44958819600287825, 0.47821048900368623, 0.5064199720000033, 0.5341635720033082, 0.5626784180058166, 0.5906799850054085, 0.6194172770046862, 0.6476894150036969, 0.6754923910048092, 0.7041253600036725, 0.732218177006871, 0.7598415180036682, 0.7882466770024621, 0.8161751720035682, 0.8448267350031529, 0.8730216870026197], '1': [0.001398479005729314, 0.028799166000680998, 0.05703140900004655, 0.0850068830040982, 0.11345360400446225, 0.1415548060031142, 0.16926023000269197, 0.1977993980035535, 0.2256220090057468, 0.25427169700560626, 0.2825470130046597, 0.31025045700516785, 0.3390184260060778, 0.3668695260057575, 0.39459536600043066, 0.42281651600205805, 0.4506980840014876, 0.4793854280069354, 0.5079920960051822, 0.5354365840030368, 0.5639478350058198, 0.5919007580014295, 0.6205333620018791, 0.6487799280002946, 0.6765702480042819, 0.7053996220056433, 0.7333150960039347, 0.7609140109998407, 0.7895249490029528, 0.8172817790036788, 0.8459176130054402, 0.8741036060018814]}
2024-02-02T00:13:30-0500 DEBUG [adc_reader] aggregated_signals={'2': [1309, 913, 1768, 1423, 797, 1580, 1568, 802, 1440, 1796, 959, 1201, 1854, 1155, 1057, 1848, 1248, 836, 1689, 1488, 799, 1518, 1734, 899, 1265, 1855, 1101, 1100, 1841, 1229, 854, 1718], '1': [1373, 1236, 1407, 1385, 1238, 1357, 1424, 1256, 1334, 1452, 1310, 1296, 1447, 1343, 1264, 1439, 1358, 1241, 1360, 1407, 1247, 1346, 1451, 1296, 1307, 1450, 1334, 1279, 1438, 1359, 1242, 1396]}
2024-02-02T00:13:31-0500 DEBUG [adc_reader] timestamps={'2': [5.385399708757177e-05, 0.02767698700336041, 0.05597745800332632, 0.08389986000111094, 0.11250230899895541, 0.1406428859991138, 0.1682979460019851, 0.19727195399900666, 0.22523107400047593, 0.25391612599923974, 0.28222342100343667, 0.3100099380026222, 0.33854660599899944, 0.36659484000119846, 0.39417797300120583, 0.4225287049994222, 0.4504077219971805, 0.4790240240035928, 0.507250226000906, 0.5350601280006231, 0.5635567959980108, 0.5915425310013234, 0.6202921139993123, 0.6485634709970327, 0.6763673320019734, 0.7049932180016185, 0.7330949410024914, 0.7607345849974081, 0.7891371389996493, 0.8170662069969694, 0.8457254259992624, 0.8739162109995959], '1': [0.0048420480015920475, 0.02894213399849832, 0.05708849200163968, 0.08519469400198432, 0.11378198699821951, 0.14178912700299406, 0.16938377099722857, 0.19851178899989463, 0.22648898199986434, 0.25501304600038566, 0.28351231799751986, 0.3112232630010112, 0.33973519100254634, 0.36767894699733006, 0.39525504899938824, 0.4237153110007057, 0.4515017759986222, 0.48012667299917666, 0.508607559997472, 0.5364765760023147, 0.5648255370033439, 0.5926659600008861, 0.6215132509969408, 0.6496545569971204, 0.6774696680004126, 0.7062868029970559, 0.7343709740016493, 0.7619903570011957, 0.79028931799985, 0.8183461980006541, 0.8468253660030314, 0.8750030779992812]}
2024-02-02T00:13:31-0500 DEBUG [adc_reader] aggregated_signals={'2': [907, 1665, 835, 1312, 1835, 1032, 1168, 1838, 1295, 837, 1630, 1532, 810, 1464, 1616, 832, 1365, 1814, 1048, 1105, 1842, 1201, 849, 1712, 1463, 805, 1466, 1620, 825, 1344, 1821, 1007], '1': [1243, 1441, 1277, 1313, 1454, 1325, 1295, 1433, 1363, 1240, 1364, 1412, 1252, 1342, 1436, 1278, 1324, 1453, 1328, 1277, 1441, 1347, 1246, 1387, 1404, 1248, 1333, 1440, 1274, 1315, 1455, 1317]}
2024-02-02T00:13:31-0500 DEBUG [adc_reader] ADC offsets: {'2': 0, '1': 0}, and in voltage: {'2': 0.0, '1': 0.0}
2024-02-02T00:13:35-0500 INFO [od_reading] Ready.
2024-02-02T00:13:35-0500 DEBUG [od_reading] od_reading is blocking until disconnected.
2024-02-02T00:13:36-0500 DEBUG [adc_reader] timestamps={'2': [4.994700429961085e-05, 0.027613600999757182, 0.055939281002792995, 0.0837846520007588, 0.11237861100380542, 0.1404964800021844, 0.16816601900063688, 0.1965737810023711, 0.22460863000014797, 0.2533413900018786, 0.2815651440032525, 0.3093248380027944, 0.3378549960034434, 0.3659162510011811, 0.3935118320005131, 0.42183495900098933, 0.44972585100185825, 0.4783769450004911, 0.5065124700049637, 0.5342065400036518, 0.5626713329984341, 0.5906593070030794, 0.6193768590019317, 0.6476485290040728, 0.6754814000014449, 0.7040274430037243, 0.7321285409998382, 0.759776465005416, 0.7882224050044897, 0.8161443900025915, 0.844824701998732, 0.8730506960055209], '1': [0.0018975380007759668, 0.028826613001001533, 0.05730062500515487, 0.08491245600453112, 0.11352620700199623, 0.1416804820037214, 0.16941371800203342, 0.19774695000523934, 0.2258786210004473, 0.25459263100492535, 0.2826819070032798, 0.3104165490003652, 0.3389556129986886, 0.36709322100068675, 0.39461359399865614, 0.4229331809983705, 0.4509145920019364, 0.47962146699865116, 0.5075910040031886, 0.5352980419993401, 0.5638686680031242, 0.5917615910002496, 0.6204606009996496, 0.6487943540050765, 0.6766949330049101, 0.705115664000914, 0.7332685330038657, 0.7609475510034827, 0.7894878639999661, 0.8172983920012484, 0.8460004750013468, 0.8742823010034044]}
2024-02-02T00:13:36-0500 DEBUG [adc_reader] aggregated_signals={'2': [1216, 1312, 1360, 1216, 1329, 1330, 1314, 1367, 1361, 1465, 1437, 1279, 1446, 1313, 1287, 1427, 1271, 1311, 1345, 1215, 1341, 1332, 1417, 1415, 1305, 1464, 1376, 1284, 1435, 1293, 1296, 1344], '1': [992, 991, 987, 989, 992, 989, 1007, 1002, 978, 1014, 980, 980, 993, 970, 986, 990, 973, 992, 988, 993, 992, 986, 1012, 990, 977, 996, 971, 986, 992, 970, 990, 989]}
2024-02-02T00:13:41-0500 DEBUG [adc_reader] timestamps={'2': [5.3385003411676735e-05, 0.02769167400401784, 0.05597860400303034, 0.08380928700353252, 0.11240225700021256, 0.1405272089978098, 0.16824424699734664, 0.19668633300170768, 0.22465243199985707, 0.2534061819969793, 0.28164014300273266, 0.30944629599980544, 0.33796228699793573, 0.3660260929973447, 0.3936086539979442, 0.4219448020012351, 0.4498207979995641, 0.4784302780026337, 0.5065959069979726, 0.5343042989989044, 0.562748989003012, 0.5907255560014164, 0.6195920660029515, 0.647858319003717, 0.6756671800030745, 0.7042263990006177, 0.7323539040007745, 0.7599955260011484, 0.7883861009977409, 0.8163007940020179, 0.8449833470003796, 0.8731952250018367], '1': [0.0013823859990225174, 0.029052600999420974, 0.05712104400299722, 0.0849205289996462, 0.11358178000227781, 0.14170975299930433, 0.16963652899721637, 0.19794476099923486, 0.2258446109990473, 0.2546812240034342, 0.28274591600347776, 0.3106481619979604, 0.33913769500213675, 0.367137127002934, 0.3946857810005895, 0.4231019290018594, 0.45091552899975795, 0.47956287400302244, 0.5077218360020197, 0.5354759050023858, 0.5638287199981278, 0.5918056529990281, 0.6207796090020565, 0.6489494569977978, 0.6768006089987466, 0.7053797240005224, 0.7335765509997145, 0.7611085910029942, 0.789576508999744, 0.8174195879983017, 0.8462008900023648, 0.8743124570028158]}
2024-02-02T00:13:41-0500 DEBUG [adc_reader] aggregated_signals={'2': [1295, 1447, 1303, 1297, 1374, 1211, 1316, 1332, 1298, 1375, 1348, 1432, 1424, 1272, 1468, 1407, 1289, 1429, 1250, 1297, 1359, 1216, 1351, 1358, 1411, 1421, 1295, 1455, 1423, 1279, 1424, 1250], '1': [984, 991, 969, 986, 981, 990, 992, 987, 1004, 993, 976, 1017, 977, 979, 1006, 971, 986, 991, 977, 989, 979, 992, 992, 980, 1017, 992, 979, 1014, 975, 980, 992, 974]}
2024-02-02T00:13:46-0500 DEBUG [adc_reader] timestamps={'2': [5.7604003814049065e-05, 0.027711310001905076, 0.056060269998852164, 0.08390782900096383, 0.1124791840047692, 0.14062226100213593, 0.16829174799931934, 0.19669888500357047, 0.2246319640034926, 0.25331217200437095, 0.28160550800384954, 0.3093704630009597, 0.3378981729983934, 0.3659678129988606, 0.39355870600411436, 0.4218823550036177, 0.44975256900215754, 0.47837402800359996, 0.5065279900009045, 0.5342371119986637, 0.5627009680029005, 0.5906864420030615, 0.6194045150041347, 0.6476612370024668, 0.6754771810010425, 0.7040216090026661, 0.7321746860034182, 0.7598105270008091, 0.7882112060033251, 0.8161306910042185, 0.8447817850028514, 0.8729999659990426], '1': [0.0014238439980545081, 0.029034842002147343, 0.057361928003956564, 0.08504641400213586, 0.11357615500310203, 0.14180730400403263, 0.16948293700261274, 0.1977889289992163, 0.22571117500046967, 0.2544110700036981, 0.28290211399871623, 0.3104537370018079, 0.3390627999979188, 0.3671630649987492, 0.3946440630024881, 0.42301177400076995, 0.4508379260005313, 0.4795437069988111, 0.5076033470031689, 0.5353759060017182, 0.5639179909994709, 0.5917781530006323, 0.6204966420045821, 0.6487373749987455, 0.6766601929994067, 0.705097329999262, 0.7333146259989007, 0.7609569769992959, 0.7893804690029356, 0.8172216730017681, 0.8458675590009079, 0.8742312070025946]}
2024-02-02T00:13:46-0500 DEBUG [adc_reader] aggregated_signals={'2': [1377, 1354, 1446, 1409, 1280, 1467, 1422, 1290, 1434, 1237, 1313, 1355, 1312, 1352, 1334, 1388, 1360, 1323, 1459, 1422, 1279, 1458, 1280, 1299, 1383, 1226, 1328, 1327, 1351, 1358, 1344, 1456], '1': [992, 976, 1014, 992, 976, 998, 976, 986, 992, 978, 991, 983, 1001, 993, 985, 1019, 994, 979, 1010, 971, 980, 994, 971, 990, 982, 993, 992, 986, 1003, 992, 975, 1011]}
2024-02-02T00:13:51-0500 DEBUG [adc_reader] timestamps={'2': [5.406199488788843e-05, 0.027743599996028934, 0.05613896699651377, 0.0840286609964096, 0.11271470199426403, 0.14088694599922746, 0.1685422139998991, 0.19696132999524707, 0.2248970129949157, 0.25357201299630105, 0.2818618079982116, 0.3096315019938629, 0.33815134699398186, 0.36620343499816954, 0.3938104219996603, 0.4221485499947448, 0.45001652499922784, 0.47862511899438687, 0.506793716995162, 0.5345056509977439, 0.5629440899938345, 0.5909425849968102, 0.6196974799968302, 0.6479705039964756, 0.6757865519975894, 0.7043623339995975, 0.7324640579972765, 0.7600946379970992, 0.7885063590001664, 0.8164277709947783, 0.845080479994067, 0.8732798579949304], '1': [0.0013682709977729246, 0.029037184998742305, 0.057665205997182056, 0.08533656799409073, 0.11404760900040856, 0.1421482909936458, 0.16964100799668813, 0.19816486299532698, 0.2259778909938177, 0.254656952994992, 0.28312940299656475, 0.3107661289977841, 0.3392805059993407, 0.3673106669957633, 0.394985569997516, 0.4233435930000269, 0.45109641299495706, 0.4797117259950028, 0.5079755829938222, 0.5356319969941978, 0.5640306969944504, 0.5921226279970142, 0.6209209089938668, 0.6490702359951683, 0.6769119089949527, 0.7055711799985147, 0.7335852989999694, 0.761181713998667, 0.7897345789970132, 0.8175254710004083, 0.846170106997306, 0.8744263599946862]}
2024-02-02T00:13:51-0500 DEBUG [adc_reader] aggregated_signals={'2': [1366, 1358, 1441, 1393, 1277, 1456, 1425, 1284, 1455, 1248, 1312, 1370, 1277, 1333, 1325, 1358, 1354, 1345, 1456, 1421, 1280, 1454, 1289, 1296, 1404, 1225, 1327, 1341, 1344, 1353, 1358, 1448], '1': [995, 981, 1011, 993, 980, 1002, 976, 980, 993, 975, 992, 984, 996, 993, 984, 1011, 993, 976, 1010, 990, 978, 994, 969, 990, 981, 992, 991, 983, 1003, 992, 978, 1017]}
^C2024-02-02T00:13:53-0500 DEBUG [od_reading] Exiting caused by signal Interrupt.
2024-02-02T00:13:53-0500 INFO [od_reading] Disconnected.
2024-02-02T00:13:53-0500 DEBUG [od_reading] Disconnected successfully from MQTT.