Reference photodiode fails self test

Hey team!

I have an error with the reference photodiode failing during self-test and am apprehensive about continuing with the calibration. Here is my error message:

self test test REF is lower than 0 dot 256 volts: Recorded 0.019 in REF, should ideally be between 0.02 and 0.500. Current IR LED: 70.0%.

Things I tried:
The leader has a different version, tried updating but failed, will post this issue as it reoccurs with a different group of pios.

Changing the ref photodiode out for a new one with no change in behaviour. I gather that this is a software issue in that case.

Changing the IR LED value to 90% in the config file offset the recorded value in REF slightly but did not eliminate the issue. Current settings following

[od_reading.config]
samples_per_second=0.2
ir_led_intensity=70
pd_reference_ema=0.9
smoothing_penalizer=6.0
turn_off_leds_during_reading=1

[od_config.photodiode_channel]
1=REF
2=90

Any assistance is appreciated! Thanks!

Hi @domatv,

Hm, that does seem low. Can you remove the IR LED, and inspect the “pocket” it sits in. Specifically, there is a small hole on the right-hand-side that leads to the REF PD. Can you make sure that doesn’t look deformed in any way?

Also, confirm that the PD’s leads aren’t touching each other.

Do any other self-tests fail or just this one?

Hey, I tested with a flashlight and can confirm there is light getting into the PD and the hole was fine.

I misunderstood how the ref PD works so I changed the IR diode and it is now passing the self-test!

Thanks!

I’m having issues with both photodiode on the self tests. I just installed a replacement the HAT on my pioreactor and since I have never had success on any self test relating to the photodiodes.

Is it possible there is something I need to do in the software since I’ve replaced the HAT?

I have double and triple check that the HAT has be installed correctly
(the stirring isn’t working since the new HAT, the “Pioreactor HAT is detected,” heating, and AUX power pass every time in the self test)

Hi @Karm.RFA,

No software changes needed. Are the PDs in the correct position on the HAT (1 → REF, 2->90 usually). Check your config.ini’s section [od_config.photodiode_channel] to confirm.

stirring isn’t working since the new HAT

What error are you seeing here?

Solved.

With help from a coworker we realized the Model drop down menu wasn’t selected and must have reset to blank while we were working with the software in replacing the HAT I guess.

Thanks Cam for the speed response

I think in the newer software versions, there is a catch for this. But thanks for letting me know the solution!

Hi all,
I am also seeing this error. All other tests pass but the “Reference photodiode is correct magnitude” test fails. Using latest software 26.2.23.

Here is the output from the logs, i get a few different messages depending on if it is the first try or if I use the “retry failed test” option…

  1. test REF is lower than 0 dot 256 volts: Too much noise in REF channel, observed 0.01439848609792651.
  2. test REF is lower than 0 dot 256 volts: Recorded 0.511 in REF, should ideally be between 0.02 and 0.500. Current IR LED: 80.0%.
  3. Failed tests: test_REF_is_lower_than_0_dot_256_volts

Here is what I have tried so far…

  1. Reboot, retry test multiple times.
  2. Ensure optical path is clear. It appears to be clear however around the REF board hole there is white filament is this supposed to be there? Have attached image below.

Many thanks,
Al

Oh I figured it out. There was a bug in this test for v1.5s. I’ll fix shortly. You can very likely carry on even with this self-test failing.

1 Like

You can try updating to the (just released) 26.2.26 release. It has some other bug fixes too, along with better self-tests.

1 Like

Great, this has fixed the failed test for me now. Impressed with the fast patch!
thanks, Al